L
LectriCare Ian
Here's a tip, look for a nice Fluke 1651.The 1651 is the cheaper 'entry level' model often bought 'just to pass the exams' so there are some really nice ones out there. It does not have the 'Auto' function or 'Rising Ramp' test found on the mid range 1652. Both of these are part of the RCD testing section and are well worth having. 'Auto mode' means that you can work through the 6 RCD tests (1/2I,I and 5I each at 0 and 180 degrees) without having to return to the tester each time. You stay at the distribution board and reset the RCD each time it trips. The test results are stored and can be scrolled through once the test is complete. The rising ramp test is useful for fault finding since it shows the actual current flowing when the RCD tripped. A 1651 can very easily be upgraded to 1652 spec.