I've just got of these to keep a check on my MFT as reading up on the subject, seems like I need to be checking monthly (in between the yearly calibration) that the MFT is still reading OK.
Now I know thathe CalCard is to check for trends and not absolute values but I was surprised at the variation in the measured value. The resistance figures ok but I thought that insulation is a bit out. Have emailed the guy and he's responded saying that they are OK. As long as the measured figures don't drift that is all that matters. Is this correct?
Range Measured value
0.25 0.28
0.5 0.53
1.0 1.03
2.0 2.03
5.0 5.01
0.5M 0.49M
1.0M 0.99M
2.0M 1.99M
10M 9.16M
20M 19.7M
Cheers Rich
Now I know thathe CalCard is to check for trends and not absolute values but I was surprised at the variation in the measured value. The resistance figures ok but I thought that insulation is a bit out. Have emailed the guy and he's responded saying that they are OK. As long as the measured figures don't drift that is all that matters. Is this correct?
Range Measured value
0.25 0.28
0.5 0.53
1.0 1.03
2.0 2.03
5.0 5.01
0.5M 0.49M
1.0M 0.99M
2.0M 1.99M
10M 9.16M
20M 19.7M
Cheers Rich