As a check of your meter it is generally recommended that a check of consistent results is undertaken on say a monthly basis.
This involves measuring low ohms, IR, EFLI, PFC, RCD on the same test box or circuit each time and recording the results.
Then the result from last month can be compared to this month and see if there is a significant change or if there is a slow drift over several months, this ensures that you are not slowly getting more and more inaccurate over time. (a lot can happen in a year).
There are dedicated check boxes that can test all the functions of an MFT, but they are fairly costly.
Decade boxes (boxes of accurate resistors with test points) are available but still costly, check boxes tend to be a bit cheaper and have resistances for testing the MFT resistances ranges.
For the RCD, EFLI, PFC tests this can be done from a dedicated socket in your workplace that is solely for this purpose and (barring network changes outside the property) should be consistent over a year.
View attachment 32461
View attachment 32462
You could use a result sheet like either of these to record your test results and compare. (I have just had to convert them to pdf as excel files will not upload, sorry)
Your competent person scheme may also want you to do this.