Has anyone experienced issues with continuity readings fluctuating on a Megger 1730 MFT when the test leads rotate/move in the instrument as is the unavoidable condition when actually using the tester?
I can null the leads but thereafter the reading with the leads connected to each other can jump to anything up to 1 ohm if the jack plug rotates.
The instrument was recently sent to Industrial Services Limited for repair and calibration and returned with a statement saying in specification, no recal required and they cleaned the contacts but alas no improvement. I've tried new leads but same problem.
I'm wondering about possible dry joints somewhere behind the test lead sockets.
 
Megger technical have said there have been "one or two" instances of this fault and have asked me to return the instrument to them which I will do. When I get it back I'll post the outcome.
 
I returned my 1552 to Megger twice, when it was returned it was fine for a few weeks before playing up again. The cost of returning it and hiring a replacement was becoming a hassle, so I finally bought a stand alone continuity / IR tester.
 

Reply to Continuity readings unstable using Megger MFT 1730 in the UK Electrical Forum area at ElectriciansForums.net

News and Offers from Sponsors

  • Article
Join us at electronica 2024 in Munich! Since 1964, electronica has been the premier event for technology enthusiasts and industry professionals...
    • Like
Replies
0
Views
265
  • Sticky
  • Article
Good to know thanks, one can never have enough places to source parts from!
Replies
4
Views
756
  • Article
OFFICIAL SPONSORS These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then...
Replies
0
Views
743

Similar threads

Having some issues with my companys fluke 1662/1663 tester. Zeroing 2 leads for continuity test we get the value 0.08 which is expected. The issue...
Replies
0
Views
252
FFS - popped it apart, found the same issue and thought I'd update... You guys were already there!
2
Replies
21
Views
2K

OFFICIAL SPONSORS

Electrical Goods - Electrical Tools - Brand Names Electrician Courses Green Electrical Goods PCB Way Electrical Goods - Electrical Tools - Brand Names Pushfit Wire Connectors Electric Underfloor Heating Electrician Courses
These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then CLICK HERE and post a thread with who you are, and we'll send you some stats etc

YOUR Unread Posts

This website was designed, optimised and is hosted by untold.media Operating under the name Untold Media since 2001.
Back
Top