Whilst performing various continuity tests my Di-Log 9083P is only showing zero 0:00. any help would be most appreciated.
Sophie
Sophie
Follow along with the video below to see how to install our site as a web app on your home screen.
Note: This feature may not be available in some browsers.
Discuss Di-Log - zero readings during continuity test in the Electrical Tools and Products area at ElectriciansForums.net
^^^ agreed... replace batteries, and fuse some times....!! un-null leads then start again!! good piece of kit just can get awkward at times..!
I've replaced the batteries as you've advised but still zero results. Any idea where would i get the fuse from?
thanks
Sophie
Hi Sophie, You require a T3.15A, 500V 1.25" HRC Ceramic Fuse such as SIBA List No. 70 065 65 similar to those also used in the Fluke 165xB series MFTs. When using the link below ensure you select the correct fuse from the drop down list before purchasing:
FLUKE 1651,1652,1653 & B TESTERS REPLACEMENT FUSES | eBay
Just ordered a couple
Thanks for the link
Hi Sophie did you manage to solve your problem? I'm having the same issue with my tester!!
Reply to Di-Log - zero readings during continuity test in the Electrical Tools and Products area at ElectriciansForums.net
Thanks for visiting ElectriciansForums.net, we hope you find the Electricians Tools you're looking for. It's free to sign up to and post a question yourself to find a tool or tool supplier either local to you, or online. Our community of electricians and electrical engineers will do their best to find the best tool supplier for you.