Hello people, can anyone clarify to me, which testers need leads to GS38. I always thought that all testers, ie low resistance ohmmeters, insulation resistance ohmmeters, earth fault loop impedance etc.. had to have leads to GS38. Then our lecturer on the 2391 course told us that , testers that did the dead tests like the low resistance ohmmeter and insulation resistance ohmmeters did not need them. However, in reading the latest examiners report of the 2391 exam, posted on this forum, in states that candidates lost marks for not putting down that insulation resistance ohmmeters need leads to GS38. I am confused. Not good when our exam is next thursday.



Cheers
 
Hi mate,

On page 81 of GN3, 3rd paragraph, states GS38 should be observed for all instruments.

good luck! with the exam, I have mine on the same day
 
Cheers spark, yes i have read that, thats why i was getting confused with what i have recently been told. Good luck to you too.
 
Ask your'e lecturer to justify his statement against what is said in GN38!What hope have you guys got when your'e lecturer can't get it right
 
Cheers,

There are a few other anomilies with the chief examiners report too.
On one part of the report it says candidates have been deducted marks for incorrectly using 16th ed terms such as "Main Equipotential Bonding" but this term is used on the PIR certificate (p101 GN3) and on the diagram fig 2.3a in GN3 p39,
nice to see some joined up thinking
 
Last edited:
I did 2391 in july and i was taught at college that GS38 leads only applied to live tests for safety reasons so i only mentioned them in questions about live tests and i passed,good luck
 
GS 38 is a HSE issued guidance sheet, so stating that they comply with any test test is correct. As I believe it says that all tests instruments must comply with it.
 
Changing the subject a bit, it says in that report about disconnection time for a 300mA RCD. Where does it say times for that in regs book, i cant find it, although i think the disconnection time is 300ms, for a BSEN 61008 and 61009.
 
Any RCD used for additional protection @ 30mA or below is tested at 1/2 *1 and 5*Idn (max 40 mS), every other RCD is only tested at 1/2 and 1*Idn , so a 300mA is only tested @ max *1, it also depends if it is a BS type 200mS, or a BSEN type 300mS, unless it is a time delay type

Edit: meant to say max disconnection times
 
Last edited:
theres usually a question about test leads on the 2391 so be aware. What i did was to head the answer with either "this is a dead test" or " this is a live test" before going on to explain how to perform the test....and dont forget to mention to "ask permition from the person ordering the test or any other interested parties" and to do a safe isolation etc as failure to mention any of these things will lose you marks guys....
Glenn.
 
Hi.
Are GS38 leads to cope with if you should test on a live, high energy circuit deliberately or otherwise and there is a fault in the instrument so fuse protected and also designed not to be miss connected or otherwise cause a short. This is all to do with the avaliable energy, and not say because it is a 1000V megger
 
Hi.
Are GS38 leads to cope with if you should test on a live, high energy circuit deliberately or otherwise and there is a fault in the instrument so fuse protected and also designed not to be miss connected or otherwise cause a short. This is all to do with the avaliable energy, and not say because it is a 1000V megger
Have i had too many tins or does that make no sense?
 
Hi.
Are GS38 leads to cope with if you should test on a live, high energy circuit deliberately or otherwise and there is a fault in the instrument so fuse protected and also designed not to be miss connected or otherwise cause a short.

Probably all of the above, hence the 500mA fused leads, finger guards and shrouds and max 4mm preferred 2mm tips, which means you cannot get the bliddy probes in
 
Probably all of the above, hence the 500mA fused leads, finger guards and shrouds and max 4mm preferred 2mm tips, which means you cannot get the bliddy probes in
yes like trying to get a Zs on a batten with a 5A connector block with lantern type prods...lol.......
 
I think it is generally accepted that a low resistance ohmmeter does not need to have GS38 test leads because it is testing at a low voltage on a dead circuit.
other testers are either testing at above 50V or on live equipment and therefore need the protection.
 
Hi guys - I'm stuck into June's test paper myself...and just got to question 22c!

I'd say the following testers 1) Earth Fault Loop Impedance Tester, RCD Tester, Phase Rotation Meter and Insulation Resistance Tester are the instruments requiring leads to GS38.

If you look at page 4 of GS38, under point 18, it states that 'for voltage detection or measurement, test leads protected by a fuse (or fuses) are recommended....' The question also carries four marks, so surely thats four testers to be stated?? I may be wrong!

Cheers
 
Hi guys - I'm stuck into June's test paper myself...and just got to question 22c!

I'd say the following testers 1) Earth Fault Loop Impedance Tester, RCD Tester, Phase Rotation Meter and Insulation Resistance Tester are the instruments requiring leads to GS38.

If you look at page 4 of GS38, under point 18, it states that 'for voltage detection or measurement, test leads protected by a fuse (or fuses) are recommended....' The question also carries four marks, so surely thats four testers to be stated?? I may be wrong!

Cheers
firstly mate if the course you are doing is the C&G 2391 then you need to put down gs38 and pick up GN3 as its that book that the course is based on and why an insulation resistance ohmmeter?...IR testing is dead testing mate.....but unless you post the question your stuck on ....no one can hely/advise you.......
 

OFFICIAL SPONSORS

Electrical Goods - Electrical Tools - Brand Names Electrician Courses Green Electrical Goods PCB Way Green 2 Go Pushfit Wire Connectors Electric Underfloor Heating Electrician Courses Heating 2 Go Electrician Workwear Supplier
These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then CLICK HERE and post a thread with who you are, and we'll send you some stats etc

Advert

Daily, weekly or monthly email

Thread Information

Title
GS38 Test leads
Prefix
N/A
Forum
Periodic Inspection Reporting & Certification
Start date
Last reply date
Replies
18

Advert

Thread statistics

Created
fredrico,
Last reply from
GLENNSPARK,
Replies
18
Views
14,120

Advert

Back
Top