Hi all,
I'm getting an odd reading from my megger 1552,when trying a zs it is showing >500 and the fuse symbol is showing.
I have got new fuses,with the same result.
Has the tester somehow got fried from greater than 500volts?
 
Hi all,
I'm getting an odd reading from my megger 1552,when trying a zs it is showing >500 and the fuse symbol is showing.
I have got new fuses,with the same result.
Has the tester somehow got fried from greater than 500volts?


Have you checked meter on another circuit or installation just incase you dont have a cpc connection.

Try it with 2 leads between L & N that should give reading if not meter probably fried.
 
Just to let you all know.
The tester had fried wire from the internal battery supply.
Fixed by Calmet in kingston upon thames,for ÂŁ165.........at Christmas.......doh.
Still,at least it ain't a new tester!
Happy new year all...
 
My m8 had the same problem about 2 months ago, the meter was well out of warranty phoned Megger they said it would be an internal fuse. As the meter was out of warranty we tried to get into the meter to change the fuse got it open a crack but to no avail so frustrating as I could see all 4 quick blow fuses just out of fingers reach. Anyway got a quote from SIS in Brooklands ÂŁ265+ vat and a 2 week wait just went out and bought a new one.
 

Reply to Is my megger banjaxed? in the Australia area at ElectriciansForums.net

News and Offers from Sponsors

  • Article
Join us at electronica 2024 in Munich! Since 1964, electronica has been the premier event for technology enthusiasts and industry professionals...
    • Like
Replies
0
Views
307
  • Sticky
  • Article
Good to know thanks, one can never have enough places to source parts from!
Replies
4
Views
828
  • Article
OFFICIAL SPONSORS These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then...
Replies
0
Views
929

Similar threads

  • Question
I've discussed this with the OP directly as we are friends, but for anyone searching in the future, the service manual suggests that this...
Replies
2
Views
880
L
  • Question
My Understanding is the 6 Months interval is IET Guidance and this wording is incorporated onto the label which is a requirement in 514.12.2 where...
Replies
9
Views
657

OFFICIAL SPONSORS

Electrical Goods - Electrical Tools - Brand Names Electrician Courses Green Electrical Goods PCB Way Electrical Goods - Electrical Tools - Brand Names Pushfit Wire Connectors Electric Underfloor Heating Electrician Courses
These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then CLICK HERE and post a thread with who you are, and we'll send you some stats etc

YOUR Unread Posts

This website was designed, optimised and is hosted by untold.media Operating under the name Untold Media since 2001.
Back
Top