sparkyork

hi all

as above im looking for the exact functional difference between these two testers as im looking at getting a 1652

thanks in advance

rich
 
I believe the 1652b has the facility to do Ra and a no trip Zs whereas the standard 1652 doesn't
 
cheers, i used to have a 1652b. i assumed the no trip zs was on the 1652 aswell :-(
 
ahh take it youve got the 1652 then, looks like its gotta be "b" series at least
Yeah i bought it new before the b series was out, i quite like the look of the 1654b but there's better things to spend £800 on at the moment :)
 
yeah i used to have a 1652b but sold it regrettably, got a megger cm500 but couldnt get on with it. shucks lol
really suprised the 1652 doesnt have no trip zs
 
no offense intended but ive just been reading an online manual for the older 1651-1653 series testers. it mentions the three wire loop testing method and does not mention anything about tripping rcds but does mention tripping rcds when doing a high current two wire test method, are you defo using a three wire setup and still tripping rcd's?

watching a 1652 on the bay and dont wanna miss out if it turns out my wires are crossed lol
 
no offense intended but ive just been reading an online manual for the older 1651-1653 series testers. it mentions the three wire loop testing method and does not mention anything about tripping rcds but does mention tripping rcds when doing a high current two wire test method, are you defo using a three wire setup and still tripping rcd's?

watching a 1652 on the bay and dont wanna miss out if it turns out my wires are crossed lol
I never said it trips rcd's all the time but i does sometimes, i assume the b series has a low current setting as well as the normal setting for such an instance
 
The older Fluke 1652 was a non-trip loop test with the three wire method as sparkyork pointed out. The original Fluke 1650 series testers were never advertised as having a high current loop test option, although it did mention it in the manual (two wire method is a "way" of carrying out a high current loop measurement).

When the Fluke 1652B was launched, the specification made clear that there were two options for loop testing, both high current and non-trip. The "B" version also added a variable RCD setting, that allows you to customise the RCD test and a pass / fail indicator for the RCD.

In addition the later "C" version then added phase rotation.

In short, the Fluke 1652 original model will perform a non-trip earth loop test, if you want a copy of the old and new manuals, drop me a PM and I can send them over for you.
 
thanks guys! is your unit a little but faulty with it tripping on loop tests?
think ill have a bid on this 1652 as im not bothered about high loop tests unless i physically just do a two wire test.
hope im doing the right thing lol
 

OFFICIAL SPONSORS

Electrical Goods - Electrical Tools - Brand Names Electrician Courses Green Electrical Goods PCB Way Green 2 Go Pushfit Wire Connectors Electric Underfloor Heating Electrician Courses Heating 2 Go Electrician Workwear Supplier
These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then CLICK HERE and post a thread with who you are, and we'll send you some stats etc

Advert

Daily, weekly or monthly email

Thread starter

Email
Joined
Time zone
Last seen

Thread Information

Title
looking for exact difference between 1652 and 1652b
Prefix
N/A
Forum
Electrical Tools and Products
Start date
Last reply date
Replies
10

Advert

Thread statistics

Created
sparkyork,
Last reply from
sparkyork,
Replies
10
Views
3,438

Advert

Back
Top