Hi all, I am pleased to be onboard with you all and looking forward to joining in with the forum activities.

My reason for my new interest in this site today is this.

I have a "Metrel Easitest" multifunction test set and ever since I had it I have never trusted the trip lock readings of Zs. For instance on an expected Zs of say 0.4 ohm I get readings variant from 0.01 to 1.9 ohm.
I have done a trial test with 20 readings, it took a long time and I then obtained an average which is very close to expected results but why should this be. Is this normal for this test set/other test sets, or a fault.

With the 17th edition trip lock is even more necessary, so should I ignore the trip lock facility and just add ,measured Ze to R1+R2, previously measured on continuity, and record the Zs result that way.
Help please
 
Last edited by a moderator:
Hi Ihave the Metrel euro test Mi3002 and on the Zsrcd setting I also get readings that do not concur with Ze + R1+R2 valies calculated from tests, it also goes in and out of calibration after nulling out leads etc maybe a Metrel thing. I intend to get onto tech Monday morn as this tester is only months old
in the meantime maybe best sticking with the formulae above that you know and love so well
Mac
 
Hi

We run all Metrel test gear and have had problems but once we voiced our disgust rather loudly with the Metrel rep one day we got two of our meters changed FOC.
Few questions. 1. Are you on Zsrcd setting? 2. Are you using three leads? 3. Are you using the tip commander? 4. Is your R1 + R2 correct due to nulling leads?

If you are on Zsrcd then it will test using a 25amp balanced test so hence the RCD will not trip (And to give Metrel their due I have yet to find a RCD it will trip)
You can only use a three lead or plugtop lead to test Zs though a RCD. Balance issue again. If you are using the tip commander then throw it away. The components in it throws the readings.

The other thing to consider is are the R1 + R2 readings correct. We had issues and ended up making up a link out socket with a 1gang trailing lead socket and the socket and see plugtop link unit. The one with the three round holes to accept the leads directly. Now we know are leads are nulling correctly rather than just holding them together we get much better readings.
 

Reply to METREL "EASITEST" Zs PROBLEMS in the Electrical Tools and Products area at ElectriciansForums.net

News and Offers from Sponsors

  • Article
As the holiday season approaches, PCBWay is thrilled to announce their Christmas & New Year Promotions! Whether you’re an engineer or an...
Replies
0
Views
775
  • Article
Bloody Hell! Wishing you a speedy recovery and hope (if) anyone else involved is ok. Ivan
    • Friendly
    • Like
Replies
13
Views
1K
  • Article
Join us at electronica 2024 in Munich! Since 1964, electronica has been the premier event for technology enthusiasts and industry professionals...
    • Like
Replies
0
Views
978

Similar threads

Duh!!! Just re-read Op's original post, it was converted to a RFC!
Replies
12
Views
795
If the shared neutral is lost ,won`t you be getting 400v across both cottages? 2 phase US style but double the Voltage.
Replies
22
Views
3K

Electricians Tools | Electrical Tools and Products

Thanks for visiting ElectriciansForums.net, we hope you find the Electricians Tools you're looking for. It's free to sign up to and post a question yourself to find a tool or tool supplier either local to you, or online. Our community of electricians and electrical engineers will do their best to find the best tool supplier for you.

We also have a Tiling Tools advice from the worlds largest Tiling community. And then the Plumbers Forums with Plumbers Tools Advice.

Search Electricans Forums by Tags

OFFICIAL SPONSORS

Electrical Goods - Electrical Tools - Brand Names Electrician Courses Green Electrical Goods PCB Way Electrical Goods - Electrical Tools - Brand Names Pushfit Wire Connectors Electric Underfloor Heating Electrician Courses
These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then CLICK HERE and post a thread with who you are, and we'll send you some stats etc

YOUR Unread Posts

This website was designed, optimised and is hosted by untold.media Operating under the name Untold Media since 2001.
Back
Top