When testing your meter with "Resistance Check Box", if you get 1.99 instead of 2 ohms, you know the reading is within the range. But what exactly is the acceptable tolerance in these cases? Is it 99%? Thanks!
 
a small change in temperature will give you more variation than that. i wouldn't worry.
 
GN3 says a basic accuracy of 5% (+- 2.5%) is usually adequate. Analogue instruments require a basic accuracy of 2% of full-scale-deflection in order to provide the required accuracy of measurement over a useful proportion of the scale.
 
GN3 says a basic accuracy of 5% (+- 2.5%) is usually adequate. Analogue instruments require a basic accuracy of 2% of full-scale-deflection in order to provide the required accuracy of measurement over a useful proportion of the scale.

That's actually ± 5% (95% - 105%).

Different checkboxes have different working tolerances. Mine is ± 5%.

The CalCard which is recommended by the Part P schemes is ± 3%.
 
I beg to differ GN3 says 5%, that's the intrinsic uncertainty. +-2.5% of nominal.
BS EN 61557, however, gives intrinsic uncertainties of:- Volts 0.8%, R[SUB]LO [/SUB]1.5%, R[SUB]INS [/SUB]10%, Z[SUB]L [/SUB]6%, Delta T 1%, I Delta n 5% and R[SUB]E [/SUB]3.5%.
The actual values of resistors in any check box is irrelevant as long as they span the range of the instrument. It's not a calibrator or standard, merely a check method. The important part of the check box routine is the record of results, because you seek to identify a significant trend of variance from your initial results that allows you to question the instrument accuracy.
 
Last edited by a moderator:

Reply to What tolerance is acceptable? in the Electrical Wiring, Theories and Regulations area at ElectriciansForums.net

News and Offers from Sponsors

  • Article
Join us at electronica 2024 in Munich! Since 1964, electronica has been the premier event for technology enthusiasts and industry professionals...
    • Like
Replies
0
Views
380
  • Sticky
  • Article
Good to know thanks, one can never have enough places to source parts from!
Replies
4
Views
956
  • Article
OFFICIAL SPONSORS These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then...
Replies
0
Views
1K

Similar threads

Ok so just done a quick test on mine. Megger mft on high gives 0.30 Robin 4120dl on 20ohm hives 0.22 Robin 4120 on 200ohm gives 0.28 Bearing in...
Replies
5
Views
710
I have an "old school" electricity monitor similar to this one...
Replies
3
Views
199

OFFICIAL SPONSORS

Electrical Goods - Electrical Tools - Brand Names Electrician Courses Green Electrical Goods PCB Way Electrical Goods - Electrical Tools - Brand Names Pushfit Wire Connectors Electric Underfloor Heating Electrician Courses
These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then CLICK HERE and post a thread with who you are, and we'll send you some stats etc

YOUR Unread Posts

This website was designed, optimised and is hosted by untold.media Operating under the name Untold Media since 2001.
Back
Top