Hi all, I am pleased to be onboard with you all and looking forward to joining in with the forums activities.

My reason for my new interest in this site today is this.

I have a "Metrel easitest" mutifunction test set and ever since I had it I have never trusted the trip lock readings of Zs. For instance on an expected Zs of say 0.4 ohm I get readings variant from 0.01 to 1.9 ohm.
I have done a test with 20 readings and then get an average which is very close but why should this be. Is this normal for this test set or a fault.

With the 17th edition trip lock is even more necessary, so should I ignore the trip lock facility and just add Ze to R1+R2, previously measured on continuity, and record the Zs result that way.
Help please.
 
hi greeny a zs reading will hardly never be the same result as a calculated one, if your meter has been calibrated
just trust the meter and use the results.
 
Greeny, I believe that there has been an issue with this tester calculating any earth loop reading on a TT supply. Not sure about on a TN system though. Maybe send an email to metrel and maybe they can inform you if it is common.
 
I have the Metrel Eurotest meter and discovered problems very early on, the problems were sorted when I found faults/bad connections on both of the test lead sets supplied with the meter

Once fixed I got consistant readings and no error messages displayed

A couple of mates of mine have had similar problems with the same meter so my problems were not an isolated case

A few minutes with a multimeter should identify if you have similar test lead faults
 
Hi, thanks for your help, did you only have Zs problems with the RCD trip lock readings.
My results on continuity, rcd tests and insulation are fine, so I presume my leads are ok.
However I will give them a go with a multimeter to be sure.
The Zs readings have always been inconsistent from new.
Thanks again
Greeny
 
The fault I found on my lead set was on the earth banana plug, I got good readings until the live tests and then had inconsistant readings on ELI on triplock and normal settings

I just stripped the plug down and resoldered the cable/plug joint worked fine ever since

You should check the other test lead as well as both myself and one of my mates have had problems with loose connections on the plug top causing errors on tests
 

Reply to Zs trip lock problem on Metrel Easitest meter in the Electrical Tools and Products area at ElectriciansForums.net

News and Offers from Sponsors

  • Article
Join us at electronica 2024 in Munich! Since 1964, electronica has been the premier event for technology enthusiasts and industry professionals...
    • Like
Replies
0
Views
507
  • Sticky
  • Article
Good to know thanks, one can never have enough places to source parts from!
Replies
4
Views
1K
  • Article
OFFICIAL SPONSORS These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then...
Replies
0
Views
2K

Similar threads

Duh!!! Just re-read Op's original post, it was converted to a RFC!
Replies
12
Views
711
  • Question
CONCLUSION (Couldn't see how to edit title) It was not belting it down with rain today, so lifted the manhole cover. The pump is about 2 metres...
2 3 4
Replies
45
Views
6K

Electricians Tools | Electrical Tools and Products

Thanks for visiting ElectriciansForums.net, we hope you find the Electricians Tools you're looking for. It's free to sign up to and post a question yourself to find a tool or tool supplier either local to you, or online. Our community of electricians and electrical engineers will do their best to find the best tool supplier for you.

We also have a Tiling Tools advice from the worlds largest Tiling community. And then the Plumbers Forums with Plumbers Tools Advice.

Search Electricans Forums by Tags

OFFICIAL SPONSORS

Electrical Goods - Electrical Tools - Brand Names Electrician Courses Green Electrical Goods PCB Way Electrical Goods - Electrical Tools - Brand Names Pushfit Wire Connectors Electric Underfloor Heating Electrician Courses
These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then CLICK HERE and post a thread with who you are, and we'll send you some stats etc

YOUR Unread Posts

This website was designed, optimised and is hosted by untold.media Operating under the name Untold Media since 2001.
Back
Top