D

dannyv

Has any one had problems with continuity readings on their fluke I keep getting 0.00 ohms on all ring tests and it's only been calibrated 3 months ago
 
Not on mine.
I got mine 2nd hand too. Only thing that has gone wrong is the test lead with the test button on its probe.

Get it calibrated, try different test leads or see if the calibration company can fix it.

Sent from my Samsung Galaxy S II via Tapatalk
 
Could be somwthing wrong with the leads or the null facility on the tester,get a drum of earth cable and do an end to end test on the drum,i have a calcard which i use to keep my tester upto date every 2 weeks,if you do this it is acceptable to get the mft done every 3 years as long as you think the tester has no problems then it should work fine and is within tolerances,most people just have them done yearly without argument but the nic/elecsa like you to also keep a fortnightly log.
 
I don't think your leads are zeroed properly.When you zero the leads do it a couple of times to make sure that the connection is good. On mine I find if I hold the leads one way or another it can make a .02 or so Ohms difference
 
Even an impropernuller of the leads won't give a 0 ohms reading of aring, unless its a super small ring final that is lol
yes of course David...which makes me wonder wether fused prods n crocs are being employed here....and the fuse in one-or more of em has blown........or it could be the interface with the tester thats at fault....or as has been suggested already..just a case of dead batterys...
 
hello i am currently having the same problem with a replacement 1652 for my stolen 1653, any update or information on this subject would be much appreciated as i have my elecsa inspection on the 23rd May

regards danny
 
hello i am currently having the same problem with a replacement 1652 for my stolen 1653, any update or information on this subject would be much appreciated as i have my elecsa inspection on the 23rd May

regards danny

Well done resurrecting this thread Danny..............just shows the professionals get confused with dead short ( zero) and open cct ( > 500) lol.

Sorry mate...........cudn't resist :)

D.Marney..............change your batteries then come back if it didn't work mate and we'll go from there.
 

Reply to Fluke 1652 in the Australia area at ElectriciansForums.net

News and Offers from Sponsors

  • Article
Join us at electronica 2024 in Munich! Since 1964, electronica has been the premier event for technology enthusiasts and industry professionals...
    • Like
Replies
0
Views
332
  • Sticky
  • Article
Good to know thanks, one can never have enough places to source parts from!
Replies
4
Views
859
  • Article
OFFICIAL SPONSORS These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then...
Replies
0
Views
983

Similar threads

Having some issues with my companys fluke 1662/1663 tester. Zeroing 2 leads for continuity test we get the value 0.08 which is expected. The issue...
Replies
0
Views
265
I have the Martindale ET4 not a bad meter, seen them on flea bay less than ÂŁ100 new
Replies
8
Views
1K

OFFICIAL SPONSORS

Electrical Goods - Electrical Tools - Brand Names Electrician Courses Green Electrical Goods PCB Way Electrical Goods - Electrical Tools - Brand Names Pushfit Wire Connectors Electric Underfloor Heating Electrician Courses
These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then CLICK HERE and post a thread with who you are, and we'll send you some stats etc

YOUR Unread Posts

This website was designed, optimised and is hosted by untold.media Operating under the name Untold Media since 2001.
Back
Top