I had err 1 on 1653, sent to fluke who updated the firmware tested it and sent it back very quick service but relieved me of 95.50 +vat.
 
I had this problem with my tester. Rang around all the calibration houses in the Newcastle upon Tyne area, as I didn't want to send it away to Fluke, there is a company called AMECal. They offer all types of repair and calibration services, I use them all the time now. They repaired and calibrated my Fluke 1652 with error 1 for ÂŁ200+vat, a lot cheaper than fluke. I also had the tester back next day, a lot faster than Fluke. There turn around time is 2-3 days, it will depend on your repair and the availability of parts. Would defiantly recommend
 
ERR1 on the Fluke 1650 series is not one fault. It simply means that the tester has failed its self-test.
There are three categories of Err1:
Non clicking Err1's testers which have lost their Calibration constants. These require a small electronic repair and long recalibration procedure.
Clicking Err1's require the same lengthy recalibration procedure without adjustment, until the fault is identified and rectified.
10A fuse Err1's. The user fits the 10A fuse from the lead probe in place of the fuse in the base of the tester. In the event of live voltage being applied to the tester whilst doing a continuity test, the fuse is unable to blow, failing to protect the circuitry. Repairs on these testers normally involves replacing 20+ blown components and in some cases these are beyond economical repair.
 
Last edited by a moderator:
ERR1 on the Fluke 1650 series is not one fault. It simply means that the tester has failed its self-test.
There are three categories of Err1:
Non clicking Err1's testers which have lost their Calibration constants. These require a small electronic repair and long recalibration procedure.
Clicking Err1's require the same lengthy recalibration procedure without adjustment, until the fault is identified and rectified.
10A fuse Err1's. The user fits the 10A fuse from the lead probe in place of the fuse in the base of the tester. In the event of live voltage being applied to the tester whilst doing a continuity test, the fuse is unable to blow, failing to protect the circuitry. Repairs on these testers normally involves replacing 20+ blown components and in some cases these are beyond economical repair.

Welcome to the forum CTUK :), Feel free to use the forum and offer advice but if you wish to promote your services then please read our rules on advertising and self promotion, if you wish to advertise your services you can contact any staff and enquire about becoming a forum sponsor.

As this thread is 4yrs old with last replay 2 yrs ago I doubt we will see any further response thus the reason I will close the thread :).
 
Last edited by a moderator:

Reply to Fluke err 1 message in the Periodic Inspection Reporting & Certification area at ElectriciansForums.net

News and Offers from Sponsors

  • Article
Join us at electronica 2024 in Munich! Since 1964, electronica has been the premier event for technology enthusiasts and industry professionals...
    • Like
Replies
0
Views
257
  • Sticky
  • Article
Good to know thanks, one can never have enough places to source parts from!
Replies
4
Views
739
  • Article
OFFICIAL SPONSORS These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then...
Replies
0
Views
721

Similar threads

Having some issues with my companys fluke 1662/1663 tester. Zeroing 2 leads for continuity test we get the value 0.08 which is expected. The issue...
Replies
0
Views
252
Hi all, The electrical network in my house (Antwerp, BE) is quite old and consists of two phases of about 140 volts. This gives me approximately...
Replies
0
Views
659

OFFICIAL SPONSORS

Electrical Goods - Electrical Tools - Brand Names Electrician Courses Green Electrical Goods PCB Way Electrical Goods - Electrical Tools - Brand Names Pushfit Wire Connectors Electric Underfloor Heating Electrician Courses
These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then CLICK HERE and post a thread with who you are, and we'll send you some stats etc

YOUR Unread Posts

This website was designed, optimised and is hosted by untold.media Operating under the name Untold Media since 2001.
Back
Top