Was testing earlier and my measured R1+R2 values decreasing the further I was from the DB...

Turns out I'd nulled with my 30m wander lead earlier, but even though the machine had been switched off it hadn't reset. It was taking every reading and subtracting the 0.50 ohms measured on the long lead. No errors thrown up by the machine, so it was displaying a 0.15 ohm continuity as 0.35 (in reality it must have read -0.35 but my MFT doesn't display a negative symbol...

Is this normal?!

It's a Metrel MI3000...
 
Was testing earlier and my measured R1+R2 values decreasing the further I was from the DB...

Turns out I'd nulled with my 30m wander lead earlier, but even though the machine had been switched off it hadn't reset. It was taking every reading and subtracting the 0.50 ohms measured on the long lead. No errors thrown up by the machine, so it was displaying a 0.15 ohm continuity as 0.35 (in reality it must have read -0.35 but my MFT doesn't display a negative symbol...

Is this normal?!

It's a Metrel MI3000...

there by lies the problem throw it away and buy a fluke that way you get proper results every time
 
Not going to be drawn in to the Fluke vs Megger vs Metrel vs anything else row!

For everyone's info, I bought the Metrel as the current version of the "industry standard" Megger didn't work properly and kept tripping RCDs!
 
Was testing earlier and my measured R1+R2 values decreasing the further I was from the DB...

Turns out I'd nulled with my 30m wander lead earlier, but even though the machine had been switched off it hadn't reset. It was taking every reading and subtracting the 0.50 ohms measured on the long lead. No errors thrown up by the machine, so it was displaying a 0.15 ohm continuity as 0.35 (in reality it must have read -0.35 but my MFT doesn't display a negative symbol...

Is this normal?!

It's a Metrel MI3000...
I would have thought this was normal, the MFT will remember the settings you have even if you have switched off.
other wise if you are doing lots of testing on a long lead you would need to have the MFT on all the time as you wandered round the building. Just need to remember to null every time you change.
 
I would have thought this was normal, the MFT will remember the settings you have even if you have switched off.
other wise if you are doing lots of testing on a long lead you would need to have the MFT on all the time as you wandered round the building. Just need to remember to null every time you change.

But to display a negative result as positive sounds a bit dodgy? Easy enough to test though - null with your long lead then remove the long lead and do a zero check. What do you get?

Anyway, my fluke is not much better, anything negative gets displayed as zero. Which is why I have a 1 ohm resistor taped to it which I can use as a sanity check when I need to.
 
Not going to be drawn in to the Fluke vs Megger vs Metrel vs anything else row!

For everyone's info, I bought the Metrel as the current version of the "industry standard" Megger didn't work properly and kept tripping RCDs!

And also doesn't display a negative symbol
 
Um.... am I reading this right? You zeroed your meter to your r2 lead, didn't re-zero it and then wondered why you got high readings? Just re-zero it with your normal probes and bob's your auntie's live-in lover.
 
Um.... am I reading this right? You zeroed your meter to your r2 lead, didn't re-zero it and then wondered why you got high readings? Just re-zero it with your normal probes and bob's your auntie's live-in lover.

That's pretty much it to be honest!

I wasn't expecting my meter to measure a negative reading on a circuit and display it as a high positive reading!
 

Reply to Is my MFT knackered or just plain stupid?! in the Electrical Tools and Products area at ElectriciansForums.net

News and Offers from Sponsors

  • Article
As the holiday season approaches, PCBWay is thrilled to announce their Christmas & New Year Promotions! Whether you’re an engineer or an...
Replies
0
Views
546
  • Article
Bloody Hell! Wishing you a speedy recovery and hope (if) anyone else involved is ok. Ivan
    • Friendly
    • Like
Replies
13
Views
990
  • Article
Join us at electronica 2024 in Munich! Since 1964, electronica has been the premier event for technology enthusiasts and industry professionals...
    • Like
Replies
0
Views
921

Electricians Tools | Electrical Tools and Products

Thanks for visiting ElectriciansForums.net, we hope you find the Electricians Tools you're looking for. It's free to sign up to and post a question yourself to find a tool or tool supplier either local to you, or online. Our community of electricians and electrical engineers will do their best to find the best tool supplier for you.

We also have a Tiling Tools advice from the worlds largest Tiling community. And then the Plumbers Forums with Plumbers Tools Advice.

Search Electricans Forums by Tags

OFFICIAL SPONSORS

Electrical Goods - Electrical Tools - Brand Names Electrician Courses Green Electrical Goods PCB Way Electrical Goods - Electrical Tools - Brand Names Pushfit Wire Connectors Electric Underfloor Heating Electrician Courses
These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then CLICK HERE and post a thread with who you are, and we'll send you some stats etc

YOUR Unread Posts

This website was designed, optimised and is hosted by untold.media Operating under the name Untold Media since 2001.
Back
Top