Probably a noddy question but here goes. When testing earth fault loop impedance do you first have to null the test instrument. If the answer is yes are there any practical tips when nulling the types of test leads that incorporate a 3 pin plug for testing socket outlets?
 
depends on which make of tester you have. with megger, you don't null for EFLI testing. i believe that with the flukes you do.
 
You need to Zero the leads on a Fluke.
It stores it memory, even after turning it off, so you only need to do it once.
Do it 1st thing in the day for peace of mind.
 
It's important to short all three leads (L,N,E) of either test lead set on a Fluke 165x series MFT, leaving say the N open circuit may cause it to introduce an error term due to the lead acting as an antenna impressing RF voltage on to the very sensitive high impedance input of the tester.
Also, although the Fluke 165x series MFT does store the current value of lead null in non-volatile memory, every time the lead set is changed (BS 1363 set to 3 lead set or vice versa) the tester MUST be re-nulled, as the MFT can't store both test lead sets null value or distinguish which is which and apply the correct null.
 
Last edited by a moderator:
I used to use an old socket but found it could give unreliable contact. I now use a 50 pence piece. Jams nicely between the plug pins! Just don't spend the 50p by mistake.... :lol:
 

Reply to Do I need to Null Test Instrument. in the Periodic Inspection Reporting & Certification area at ElectriciansForums.net

News and Offers from Sponsors

  • Article
Join us at electronica 2024 in Munich! Since 1964, electronica has been the premier event for technology enthusiasts and industry professionals...
    • Like
Replies
0
Views
303
  • Sticky
  • Article
Good to know thanks, one can never have enough places to source parts from!
Replies
4
Views
819
  • Article
OFFICIAL SPONSORS These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then...
Replies
0
Views
904

Similar threads

FFS - popped it apart, found the same issue and thought I'd update... You guys were already there!
2
Replies
21
Views
2K
  • Question
He is after the four options for 1 and 2 of the multiple choice exam, probably optimistic someone will know this. Your question 3 in incomplete to...
Replies
2
Views
1K

OFFICIAL SPONSORS

Electrical Goods - Electrical Tools - Brand Names Electrician Courses Green Electrical Goods PCB Way Electrical Goods - Electrical Tools - Brand Names Pushfit Wire Connectors Electric Underfloor Heating Electrician Courses
These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then CLICK HERE and post a thread with who you are, and we'll send you some stats etc

YOUR Unread Posts

This website was designed, optimised and is hosted by untold.media Operating under the name Untold Media since 2001.
Back
Top