I suspect my Megger 1710 tester is faulty
So I cut exactly 1 meter of 2.5 -1.5

I have nulled the probes as per the instructions.

Continuity test.

End to end
L = 0.11 Ohms
N = 0.11 "
CPC - 0.11 "

Question is this test valid.
Why is the CPC not 1.67 times 0.18 Ohms

What could I do to prove it's faulty or not before sending it off.

ADDED new duracells put in

Thanks in advance of any help.
 
Last edited by a moderator:
all those readings are far too high for 1m . try a different set of leads for starters and make sure the plugs/sockets on leads and meter are clean. another tip, when reading such low values, don't null the leads. subtract from the result after.
 
Last edited:
Thanks for the tips - the meter is virtually brand new & clean I got it 11months ago used it 3 times.
Bought the kewtech light testing KIT and that didn't work, it doesn't work with the megger supplied plug tester
or the kewtech one, I put that down to incompatiblity.

Gonna send it back before the warrantly runs out. I'm guessing I'll ring megger on monday to find their off till after xmas :sad_smile:

I work as an electricains labourer and only 4 weeks of that was domestic in the past year.
No testing whatsover.
Need to get it sorted as I have guranteed min 3 weeks domestic with a local firm.
Only got this cos I was the one who got the job.
 
my lightmates don't work with megger test leads either. you need to get it off to megger asap, get it fixed under warranty.
 
i had a 1552 that was the same. all other functions were spot on, but continuity was hit and miss, mostly miss.
 
Posted to Megger, it was signed for on 24/12/15 @ 9.02am they say 5 working days for return. :D
Lucky I noticed it cos it was bought on 12 jan 2015 and I didn't register for the 3yr warranty.
 
Yes! Got it back today the service report says:
Faulty current circuit.

They've put a date sticker on it saying it's been calibrated....result.
Bit miffed I have a repaired tester, would I have a resonable shout on insisted in a new replacement ?
 
yes! Got it back today the service report says:
Faulty current circuit.

They've put a date sticker on it saying it's been calibrated....result.
Bit miffed i have a repaired tester, would i have a resonable shout on insisted in a new replacement ?
the termsof your guarantee would be standard.manufacturer repairs or replaces athis discretion.
 

Reply to megger 1710 problems or not in the Electrical Course Trainees Only area at ElectriciansForums.net

News and Offers from Sponsors

  • Article
Join us at electronica 2024 in Munich! Since 1964, electronica has been the premier event for technology enthusiasts and industry professionals...
    • Like
Replies
0
Views
361
  • Sticky
  • Article
Good to know thanks, one can never have enough places to source parts from!
Replies
4
Views
918
  • Article
OFFICIAL SPONSORS These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then...
Replies
0
Views
1K

OFFICIAL SPONSORS

Electrical Goods - Electrical Tools - Brand Names Electrician Courses Green Electrical Goods PCB Way Electrical Goods - Electrical Tools - Brand Names Pushfit Wire Connectors Electric Underfloor Heating Electrician Courses
These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then CLICK HERE and post a thread with who you are, and we'll send you some stats etc

YOUR Unread Posts

This website was designed, optimised and is hosted by untold.media Operating under the name Untold Media since 2001.
Back
Top