N
newtothis
I suspect my Megger 1710 tester is faulty
So I cut exactly 1 meter of 2.5 -1.5
I have nulled the probes as per the instructions.
Continuity test.
End to end
L = 0.11 Ohms
N = 0.11 "
CPC - 0.11 "
Question is this test valid.
Why is the CPC not 1.67 times 0.18 Ohms
What could I do to prove it's faulty or not before sending it off.
ADDED new duracells put in
Thanks in advance of any help.
So I cut exactly 1 meter of 2.5 -1.5
I have nulled the probes as per the instructions.
Continuity test.
End to end
L = 0.11 Ohms
N = 0.11 "
CPC - 0.11 "
Question is this test valid.
Why is the CPC not 1.67 times 0.18 Ohms
What could I do to prove it's faulty or not before sending it off.
ADDED new duracells put in
Thanks in advance of any help.