Hi any advice would be nic
There's 2 options in Loop
L-PE with a sub category 3Lo, 2Hi and 2Lo
and L-L/L-N which is a 2Hi test
Here is what I think are the correct settings
Ze - L-PE on 2Hi
PSCC - L-N on 2Hi
PEFC - L-PE 2Hi same as Ze
Zs - L-PE on 3Lo
any help would be appreciated x
Thanks guys i think i got the hang of it now, your help is much appreciated. The duel display actually shows Ze & PFC at same time, there is no 'PFC' setting as such, so testing for PEFC I have to do a Ze test but with the earthing conductor connected to the met, ignore the Ze result and just look at the PFC reading. Same with PSCC do a Ze between L-N incoming and read the PFC result
I think Megger tried to be too clever rushing out the 1700 series with all its new tricks their 1500 series worked so why change it
all of the meters 1500 series do not show loop/PFC measurement though which were standard issue for everyone in my classes. So it's new to me when they combined it, personally i preferred a different setting but there you go.Most meters show loop/PFC on the same setting...
I think Megger tried to be too clever rushing out the 1700 series with all its new tricks their 1500 series worked so why change it
all of the meters 1500 series do not show loop/PFC measurement though which were standard issue for everyone in my classes. So it's new to me when they combined it, personally i preferred a different setting but there you go.
yes oldtimer I agree it wasn't broke so why try and fix it!