Twitchboy

My company does not allow us to use our own voltage/continuity testers. Instead we must use the ones they provide for us. They're pretty cheap but do the job well enough.

I have had my testers for just over a year and the label says that I shouldn't use them after that time so I handed them back to my supervisor expecting him to order me a new set as sending them off to be calibrated would cost more than purchasing new testers.

Instead, he appeared ten minutes later with the testers I had just given him with a new label and a new date. He said "there you are, they're now calibrated".

Is it possible that he could have calibrated them? What (if any) equipment would he require to have calibrated them? I have my doubts that he did anything at all other than stick a new label on the old testers.
 
Last edited by a moderator:
It may be that as it's just continuity and voltage indication it's just a check not a calibration. The device is a guide, it's not like you are writing the results on an EICR or anything like that. If it's just for indication I would say they have 'tested' it on a check bos of some description with a voltage indication check.

If the instrument was used for producing legally binding reports then proper 'calibration' would be required but as it's just an indicator I think an annual 'check' would be OK.

That would make sense if it werent for the fact that I'm perfectly capable and qualified of checking my own testers. I will need to subtly investigate this matter further. Does anyone know if these test leads have fuses in them? There doesn't seem to be. There should be shouldn't there if they are GS38?
 
[FONT=&amp]Approved 2 pole voltage testers (AVT) that indicate voltage in broad bands do not require calibrating. If the AVT incorporates a meter it requires annual calibration, just like any other meter to ensure accuracy[/FONT].

2 pole AVTs that rely on an illuminated bulb (eg. Drummond and similar types) are required to be GS38 protected by a HRC fuse (500mA) or a current limiting resistor and fuse, these are usually housed in the probes.

2 pole AVTs of the type that have a LED/LCD voltage detector with integral test probe and an interconnected second test probe, have internal GS38 protective impedance and current limiting resistors built into the detector limiting the current (usually <20mA) and energy input to the detector. Therefore they do not require an additional HRC fuse and current limiting resistor in the test leads.
 
Last edited by a moderator:
Voltage/Continuity Tester, Electronic, Voltage 0 to 480 AC, Voltage 0 to 690 DC, Calibration Frequency 50/60 Hz, Safety Rating CAT IV 600 V, Display LCD with Bargraph, Audible Continuity, Battery Type AAA, Agency Compliance CE, IP64, Auto Ranging, Special Features Low Impedance Test to Check for Ghost Voltages, 3-Phase Rotation Test, Built-in Flashlight, Double Molded Housing, Includes Batteries, Protective Cap, Pouch Case and Instructions.
 
I'd always assumed that they didn't but on checking the Operating Instruction for my VCT-GS38, which has bog standard LED indicators, it states it "must be calibrated periodically and checked by our service department at regular intervals .... recommended calibration interval of one year"
 
[FONT=&amp]Approved 2 pole voltage testers (AVT) that indicate voltage in broad bands do not require calibrating. If the AVT incorporates a meter it requires annual calibration, just like any other meter to ensure accuracy[/FONT].

What does testing voltage in broadbands mean? I googled it but I just get a load of stuff about broadband internet. When you say it incorporates a meter do you mean a digital/LCD/Anologue display of some sort as opposed to a row of LED's that indicate voltage?

2 pole AVTs that rely on an illuminated bulb (eg. Drummond and similar types) are required to be GS38 protected by a HRC fuse (500mA) or a current limiting resistor and fuse, these are usually housed in the probes.
Would these HRC fuses be removeable/replaceable? There is no way of telling whether or not my test leads have them inside and if they were in there and blew, how would I replace the fuses as the probes are a totally sealed unit?

2 pole AVTs of the type that have a LED/LCD voltage detector with integral test probe and an interconnected second test probe, have internal GS38 protective impedance and current limiting resistors built into the detector limiting the current (usually <20mA) and energy input to the detector. Therefore they do not require an additional HRC fuse and current limiting resistor in the test leads.
Fascinating stuff, are these resistors in the unit itself rather than in the probes? How would I know if my testers fall into this category? There is nothing on the data sheet that tells me if this is the case.
 
"Digital display" is not the point. I have antique needle-meters with antique calibration stickers.

"Calibrate" to what accuracy? What are you measuring?

For most of my work, I stick it on a 1.5V battery and I stick it in the wall. A fresh carbon battery reads 1.56V. I usually know my wall voltage from several meters and also lamp life/brightness, and if a meter reads close to expected that's good enough for anything I do.

"Continuity" is for trailer lamps. If it beeps when shorted, that's something. There's always a go/no-go threshold, often 100 ohms.... but even in trailer-lamps a 99 ohm "beep" is not good enough to light the lights. So when it really matters you go over to an Ohms range, and check that against a box of cheap resistors (they are all 5% now except when they are 2%).

FWIW, I have un-trimmable meters 30 years old which still read within 2% of what I think they should. Mostly they work "right", they work "odd" because the battery is weak, or they work "way wrong" because something inside is sick.

Yes, meters can drift, and old needle-meters weaken with age. My 1981 MM now reads 12.3V on what I think is a 12.6V car battery. I know that and accept it. Mostly I'm looking for 11V (sick battery) or 15V (sick regulator), not part-Volt differences. And my other meters have shown negigible drift.

If you are trimming electronic amplifiers for 0.234V at TP13, then you need to check near that 0.234V point against a better meter which is occasionally compared to an even better standard.

If you want to know if there is near 230V at a motor, sticking the meter into a familiar wall outlet may be sufficient for your company's purpose. If they want you to check for contacts opening and closing, a beep test may be good enough.

They DO need to watch for field-meters which have been dropped, abused with 8,000V, or left out in the rain to corrode. (However I'd want a weekly quick-test, not annual.)

If you must use the boss's tool, then it is the boss's problem, NOT yours.
 
Fluke T100? There's nothing to calibrate on it. You can prove it works with a proving unit (or poking the probes on L & N), takes about 10 seconds. MFT's are what get calibrated.

Use them for basic continuity (nice audible way to check ring continuity quickly), safe isolation and polarity.
 
Last edited by a moderator:
this is old post but I just came across.
Any testing instrument may require to be calibrated by the manufacturer. The instrument manual may have that recommendation as well as the duration recommended by the manufacturer to perform the calibration. Please see attached snapshot from the FLUKE T100 Voltage/Continuity Tester manual.
Any calibration of such usually conducted by the instrument manufacturer or by a well established and certified calibration house. If the company that is using the instrument decided to perform the calibration in house and by their staff then a procedure need to be stablished showing the work instructions, duration and the equipment used in the process as well as what the procedure is covering. i.e. which measurement tools are covered by this procedure.
 

Attachments

  • [ElectriciansForums.net] What exactly is required to calibrate a voltage continuity tester?
    A snapshot from the FLUKE T100 Voltage_Continuity Tester manual.png
    115.6 KB · Views: 16
Last edited:

Reply to What exactly is required to calibrate a voltage continuity tester? in the UK Electrical Forum area at ElectriciansForums.net

Similar threads

I have the Kewtech that I bought in an emergency when my trusty Fluke that lasted many years gave up. It works! It has single pole testing too btw...
Replies
3
Views
1K
Just thinking about selling MFT's and single test equipment. The idea being I buy stuff from Ebay and sell it on ebay, which I am doing at the...
    • Like
    • Friendly
Replies
0
Views
619

Recommended Sponsor News

Quickwire Now Sponsor ElectriciansForums.net

I am sure you will join me in welcoming our newest sponsors of ElectriciansForums.net - Quickwire! They have decided to join us after seeing you guys discussing their products here. Now we have an expert on board that you can chat to @Quickwire-Sam who will be happy to answer your queries!

[ElectriciansForums.net] Quickwire Now Sponsor ElectriciansForums.net


@Quickwire-Sam said "At Quickwire, we're all about speed, safety, and reliability. We're a family-run business, and every Quickwire connector is proudly made right here in the UK, ensuring noticeable quality. In short, Quickwire is the fastest connection method on the market. Whether you're an experienced electrician or a DIY enthusiast, Quickwire makes electrical connections ridiculously quick and easy. We're passionate about blending British craftsmanship with innovation. If you have any questions or if there's any way we can help, please just ask me!"

Exclusive Forum Offer! Free euro 2024 wall chart for first 10 responses!

Hi everyone,

We have 10 exclusive Uheat EURO 2024 Wallcharts to giveaway for Electricians Forum Members! The first 10 people to reply YES to this thread - I will message and get them sent out to you just in time for this years tournament! GO GO GO

PCBWay Now Sponsor ElectriciansForums.net!!

I hope you will all join me in welcoming our newest sponsor to the forum PCBWay! You can contact their friendly people by sending a message to @PCBWay who will be happy to answer your queries.

Welcome to the community and thanks for your sponsorship!

[ElectriciansForums.net] PCBWay Now Sponsor ElectriciansForums.net!!

PCBWay provides services including PCB Prototype and batch Production, PCB Assembly (SMT), 3D Printing, CNC Machining, PCB Design, Electronics Modules Selling, etc. We are committed to meeting the needs of global makers from different industries in terms of quality, delivery, cost-effectiveness, and any other demanding requests in electronics.

[ElectriciansForums.net] PCBWay Now Sponsor ElectriciansForums.net!!

OFFICIAL SPONSORS

Electrical Goods - Electrical Tools - Brand Names Electrician Courses Green Electrical Goods PCB Way Electrical Goods - Electrical Tools - Brand Names Pushfit Wire Connectors Electric Underfloor Heating Electrician Courses
These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then CLICK HERE and post a thread with who you are, and we'll send you some stats etc

YOUR Unread Posts

This website was designed, optimised and is hosted by untold.media Operating under the name Untold Media since 2001.
Back
Top