Was testing earlier and my measured R1+R2 values decreasing the further I was from the DB...

Turns out I'd nulled with my 30m wander lead earlier, but even though the machine had been switched off it hadn't reset. It was taking every reading and subtracting the 0.50 ohms measured on the long lead. No errors thrown up by the machine, so it was displaying a 0.15 ohm continuity as 0.35 (in reality it must have read -0.35 but my MFT doesn't display a negative symbol...

Is this normal?!

It's a Metrel MI3000...
 
Was testing earlier and my measured R1+R2 values decreasing the further I was from the DB...

Turns out I'd nulled with my 30m wander lead earlier, but even though the machine had been switched off it hadn't reset. It was taking every reading and subtracting the 0.50 ohms measured on the long lead. No errors thrown up by the machine, so it was displaying a 0.15 ohm continuity as 0.35 (in reality it must have read -0.35 but my MFT doesn't display a negative symbol...

Is this normal?!

It's a Metrel MI3000...

there by lies the problem throw it away and buy a fluke that way you get proper results every time
 
Not going to be drawn in to the Fluke vs Megger vs Metrel vs anything else row!

For everyone's info, I bought the Metrel as the current version of the "industry standard" Megger didn't work properly and kept tripping RCDs!
 
Was testing earlier and my measured R1+R2 values decreasing the further I was from the DB...

Turns out I'd nulled with my 30m wander lead earlier, but even though the machine had been switched off it hadn't reset. It was taking every reading and subtracting the 0.50 ohms measured on the long lead. No errors thrown up by the machine, so it was displaying a 0.15 ohm continuity as 0.35 (in reality it must have read -0.35 but my MFT doesn't display a negative symbol...

Is this normal?!

It's a Metrel MI3000...
I would have thought this was normal, the MFT will remember the settings you have even if you have switched off.
other wise if you are doing lots of testing on a long lead you would need to have the MFT on all the time as you wandered round the building. Just need to remember to null every time you change.
 
I would have thought this was normal, the MFT will remember the settings you have even if you have switched off.
other wise if you are doing lots of testing on a long lead you would need to have the MFT on all the time as you wandered round the building. Just need to remember to null every time you change.

But to display a negative result as positive sounds a bit dodgy? Easy enough to test though - null with your long lead then remove the long lead and do a zero check. What do you get?

Anyway, my fluke is not much better, anything negative gets displayed as zero. Which is why I have a 1 ohm resistor taped to it which I can use as a sanity check when I need to.
 
Not going to be drawn in to the Fluke vs Megger vs Metrel vs anything else row!

For everyone's info, I bought the Metrel as the current version of the "industry standard" Megger didn't work properly and kept tripping RCDs!

And also doesn't display a negative symbol
 
Um.... am I reading this right? You zeroed your meter to your r2 lead, didn't re-zero it and then wondered why you got high readings? Just re-zero it with your normal probes and bob's your auntie's live-in lover.
 
Um.... am I reading this right? You zeroed your meter to your r2 lead, didn't re-zero it and then wondered why you got high readings? Just re-zero it with your normal probes and bob's your auntie's live-in lover.

That's pretty much it to be honest!

I wasn't expecting my meter to measure a negative reading on a circuit and display it as a high positive reading!
 

Reply to Is my MFT knackered or just plain stupid?! in the Electrical Tools and Products area at ElectriciansForums.net

Similar Threads

  • Question
I know you’ve all be waiting in suspense for the next exciting update in this story. I promise this one will be much shorter! The OH was back at...
Replies
21
Views
4K
D
Crikey this didn't take long to deviate substantially... Thanks for replies so far. Case not really something I'm worried about, currently using...
Replies
26
Views
3K
dg2908
D

Electricians Tools | Electrical Tools and Products

Thanks for visiting ElectriciansForums.net, we hope you find the Electricians Tools you're looking for. It's free to sign up to and post a question yourself to find a tool or tool supplier either local to you, or online. Our community of electricians and electrical engineers will do their best to find the best tool supplier for you.

We also have a Tiling Tools advice from the worlds largest Tiling community. And then the Plumbers Forums with Plumbers Tools Advice.

OFFICIAL SPONSORS

Electrical Goods - Electrical Tools - Brand Names Electrician Courses Green Electrical Goods PCB Way Electrical Goods - Electrical Tools - Brand Names Pushfit Wire Connectors Electric Underfloor Heating Electrician Courses
These Official Forum Sponsors May Provide Discounts to Regular Forum Members - If you would like to sponsor us then CLICK HERE and post a thread with who you are, and we'll send you some stats etc

Trending Electricians Tools

New job, new tester; Megger 1741+, those that have one; anything good/bad to report? Work also trusting me with a Flir E5 magical heat-reading...
Replies
210
Views
95K
Absolutely correct they hold water, shrouds are totally unnecessary. 20s is 24mm spanner, 20 large is 30mm spanner.
2
Replies
41
Views
45K
If that little thing, plugs your butt, you better eat slowly. 🤣
    • Like
2 3 4 5 6 7 8 9 10 11 12 13 14
Replies
336
Views
41K
B
I'd say everyone has knowledge gaps in their area of electrical work I know i have quite a few i work on to fill them in
2 3 4
Replies
75
Views
29K
I too have been tempted.. I’m buying for my son. He’s in his2nd year at collage and looking for apprentiship. Sadly it’s almost impossible in my...
Replies
11
Views
25K
Of course. “Ye canna change the laws of physics” EDIT. oh blimey, I just spotted the date of the OP. ? Thank you very much @Lord lee jan You...
Replies
40
Views
23K
Only way to do it in a tight space
2
Replies
33
Views
21K
The display on the Mft pro is not very durable mine is broken already 😡
2
Replies
29
Views
21K
Only lawyers , solicitors and dodgy chat up lines charge for picking up the telephone 🙃
2 3 4
Replies
97
Views
20K
J
I'm not entirely certain as it's very tough, but brittle when cold.
2 3
Replies
65
Views
19K
nicebutdim
N

YOUR Unread Posts

This website was designed, optimised and is hosted by untold.media Operating under the name Untold Media since 2001.
Back
Top
AdBlock Detected

We get it, advertisements are annoying!

Sure, ad-blocking software does a great job at blocking ads, but it also blocks useful features of our website. For the best site experience please disable your AdBlocker.

I've Disabled AdBlock    No Thanks